Pressure increase after injecting!

Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.

2 posts Page 1 of 1
Dear all,
I am using Agilent 1100 with autosampler, my column is Rp8 -150x4.6mm, 5mcm. The pressure increases from 81 bar to 95 bar when needle goes to needle seat; decreases back to 81 bar during run and increases again after each injection. My sample is aqueous with pH 9.5 when my mobile phase is ACN: 25mM acetate buffer pH 4 (80:20); I know that is big difference of pH but I don't have other choices (my sample is after derivatization). Is it the reason for increase of pH? What I have to do to overcome this difference?
Thanks inadvance!

Sounds like a viscosity effect, water should be more viscous than your 80%ACN, apparently your injection amt is enough to show this. However, if the pressure is only high during the injection position it may be that you have a restriction in the injection path. If you know what the cause is the remedy is obvious.
2 posts Page 1 of 1

Who is online

In total there are 23 users online :: 1 registered, 0 hidden and 22 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Bing [Bot] and 22 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry