How to set the standard solution in a related subst method?

Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.

2 posts Page 1 of 1
I would like to know which is the best approach in choosing the concentration of the standard solution for a HPLC related substances method to use in routine. Normally after performing linearity, range and accuracy during validation, in routine a single point calibration is chosen (if intercept in linearity resulted negligible) or a 3 points calibration. We usually set the concentration of the single point calibration solution at the specification limit of the impurity (or, in the same way, we fix the central point of the 3 points calibration solutions at the specification limit of the impurity). I know other companies that set the concentration of the standard solution at a value < o > of the specification limit. Which is the correct approach?
Set the concentration of the RS standard solution at your specification limit. The standard deviation/error will be about the same as an OOS sample that is borderline.
2 posts Page 1 of 1

Who is online

In total there are 20 users online :: 0 registered, 0 hidden and 20 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: No registered users and 20 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry