The absolute error is relatively constant (+0.1%) but when we inject the unspiked sample there is absolutely none of this impurity in it. Also the blanks before and after does not show any carry over.
The peak area of the main peak in the unspiked sample is about 45000000 mVs, in the spiked experiments only about 40000000 mVs (!). The peak area of the impurity in the spiked sample is about 49000 mVs (0.3%), 70000 (0.5%) and 102000 (0.8%) while when I inject the impurity alone at the given concentrations, I get peak areas of 38000, 64000 and 105000 mVs.
Shouldn't the peak area stay the same at certain concentration, no matter what else is injected?
The problem is reproducible with different columns, eluents and sample solutions.
The product is stable under the HPLC conditions.
The RRT of the impurity is 0.94, the peaks are baseline separated. Is it possible that although baseline separation, the main peak may influence the impurity peak?
Do you have any idea what has happened here?