Shimadzu RID-10A "ALAM" Issue

Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.

5 posts Page 1 of 1
I am using a Shimadzu HPLC with a RID-10A detector to analyze sugars using an ion-exchange column and water as a mobile phase. A week ago I tried running standard ethanol samples for a calibration curve and noticed that my baseline was fluctuating greatly with an upward drift even after extensive flushing of the reference and sample cells with my mobile phase (H2O). I also balanced the cells.
I read through the user manual and determined that the source of my baseline problems were probably due to cell dirtiness. I only use water and sugars in my HPLC analysis and therefore there might have been residual sugars left in the cell that left a film. Anyways, I attempted the cleaning procedure outlined in the user manual, which was 50 mL acetone , 100 mL H2O, and 50 mL of 0.1M nitric acid followed by extensive purging of my flow lines with H2O (i.e., left overnight). My flow rate was 0.500 mL/min and the pressure was consistent. After refilling the reference cell and balancing, there was a new issue.
Instead of baseline issues, the RID LED screen now displays the message "ALAM", which pertains to the Total Energy value of ~1600, when it should be from 6000-9000. This means that there is not enough light reaching the cell. The lamp has only been run for ~2400 hours and should not have to be replaced until about 20,000 hours. I also increased the Lamp Voltage from 3.89 to 5.00, which is the maximum recommended voltage.

Currently, I am trying another cleaning method using IPA (isopropyl alcohol), Ethanol:Chloroform (80:20), and IPA again followed by extensive flushing with H2O.
My flow rate again is 0.500 mL/min at room temperature. The cell temperature has always been 40 C.

Is there anything else I should try if this does not work, before ordering a new lamp?
hi, I have a same problem with you, how do you solve the problem,.
You may check the tubing connections to solenoid valves , which connected to reference and sample cells.
The fingertight fittings sometimes getting loose and leaking , this may cause baseline and energy problems.
You may check the tubing connections to solenoid valves , which connected to reference and sample cells.
The fingertight fittings sometimes getting loose and leaking , this may cause baseline and energy problems.
thanks for your help,i used flush the system with H2O for a week, and the system performs well,but i replaced the mobile phase(H20) with THF , the "alarm" appear again.
Hei! I am having the same issue. And a non-functioning solenoid valve would actually explain very well the things I am observing. Can you provide instructions on how to check the solenoid valve? Do I have to dissassemble the detector? How do you by-pass the solenoid valve?

Thanks in advance,
Anna
5 posts Page 1 of 1

Who is online

In total there are 22 users online :: 1 registered, 0 hidden and 21 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: John Guajardo and 21 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry