I have some problems with the quantification of AB by FPLC.
I used the following program
- 0.5 ml/min.
- Equilibrate: 2 CV 2xPBS.
- Wash: 2 CV 2xPBS
5 CV 100 mM Citrat pH 5.0
- Elution: 5 SV 100 mM Citrat pH 2.8
- Reequilibrate: 15 SV 2xPBS
After the elution step the baseline decreases down to round about 5 mAU, before the elution step it is at 0 mAU. Therefore it is hard to quantitate if my sample peak is very low.
Do somebody know how i can eléminate this problem?

Try to inject your mobile phase & see that profile. I t can give you idea & you can play with molar ratio too.I saw this type of problem with citrate buffer not on FPLC.

A colleague already tried this before, and you also saw this problem after the elution citrat buffer was injected (a run without a sample), but this problem doesn´t occur if you only inject the wash citratbuffer. They only distinguish in the pH value!? But i need a pH of 2.8 to elute the antibodies...Some other proposals?
3 posts Page 1 of 1

Who is online

In total there are 22 users online :: 1 registered, 0 hidden and 21 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: John Guajardo and 21 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry