Can I leave the GC at 80 degrees without carrier gas flow?
Discussions about GC and other "gas phase" separation techniques.
11 posts
Page 1 of 1
The GC running in this lab has a thermal conductivity detector and HayeSep D column that is used to seperate natural gases at room temperature. However, it can be used at high temperatures and we use it at about 80 degrees celcius. My question is can we leave the GC on at this temperature without the carrier gas(He) running through as it takes two hours for the GC to heat up and it would save us alot of time if we could do this. However, we will not do it at the expense of the column. If anyone is able to answer this question, I would appreciate it very much.
-
- Posts: 5433
- Joined: Thu Oct 13, 2005 2:29 pm
Why bother turning the carrier gas off ?. Calculate the cost of the gas - depending on where you are it will come to less than 0.05 US cents per minute.
Peter
Peter
Peter Apps
-
- Posts: 3210
- Joined: Thu Sep 02, 2004 7:28 pm
Peter is quite correct, the cost of the gas is so small why bother.
It sounds like a supervisor wants to write a report how is he saving the company money by doing everything possible.
In any case if you have a steel column not glass you could disconnect the column from the detector and put a cap on the end of the column while keeping the pressure low on the column to reduce useage,
or simply lower the gas flow down to 1 or 2 cc/min.
The cost of the cap should be returned in 1 or 2 years and the other alternative is very inexpensive as it only costs the labor involved to set the flow and then reset the flow, but what would the actual savings be?
If labor is even 20 cents a minute, is it worth waiting 10 or 20 minutes for the flow to equilibrate if you have a sample to run?
best wishes,
Rod
It sounds like a supervisor wants to write a report how is he saving the company money by doing everything possible.
In any case if you have a steel column not glass you could disconnect the column from the detector and put a cap on the end of the column while keeping the pressure low on the column to reduce useage,
or simply lower the gas flow down to 1 or 2 cc/min.
The cost of the cap should be returned in 1 or 2 years and the other alternative is very inexpensive as it only costs the labor involved to set the flow and then reset the flow, but what would the actual savings be?
If labor is even 20 cents a minute, is it worth waiting 10 or 20 minutes for the flow to equilibrate if you have a sample to run?
best wishes,
Rod
-
- Posts: 66
- Joined: Mon Mar 06, 2006 6:37 pm
Also, you might consider a cheaper grade of helium. Unless you're looking for concentrations less than 10 ppm (which would be unusual in natural gas analysis) a "Zero" or 4.8 grade would be sufficient, and cheaper than the "UHP" or 5.0 grade you're probably using.Why bother turning the carrier gas off ?. Calculate the cost of the gas - depending on where you are it will come to less than 0.05 US cents per minute.
Or is the issue how often you have to change out the carrier gas cylinder?
Michael J. Freeman
Belle Chasse, LA
Belle Chasse, LA
-
- Posts: 1233
- Joined: Tue Oct 11, 2005 7:20 am
The column will degrade without the carrier gas at 80C, as it's a porous polymer, and the equilibrium time will be even longer once carrier is restarted. You will still have to recondition the column, and column life will decrease from years to months.
Because, unfortunately, bean counter laboratory managers are not allowed to be killed by their staff, one alternative solution is to add a bypass to the instrument front end carrier gas line.
The system usually consisting of an on-off toggle valve and a restriction that reduces the flow to 1-2 ml/min that is plumbed across the usual carrier on-off valve. You open the bypass, and close the normal valve. It can all be combined in a multi-port valve, if the savings are justified.
Why not go the whole hog, and suggest a $2,000+ time-controlled, auto-switching, multiport valve to save $100 of gas per year?. It should be worth some sort of performance bonus
.
I'd also endorse using the most appropriate grade of carrier gas for TCD and FID. I used welding grade argon and technical grade hydrogen and helium on packed and capillary column GCs with gas purifiers that were regenerated once a year. In general, most carrier gas disappears through leaks or high split vents when instruments aren't running, and using cheap gases meant instruments were always ready to go.
You need ultrapure gases if you are looking for trace impurities or have oxygen-sensitive columns or detectors. If you are buying ultra-pure gases, invest in a leak detector and some simple bypass valve setups.
Bruce Hamilton
Because, unfortunately, bean counter laboratory managers are not allowed to be killed by their staff, one alternative solution is to add a bypass to the instrument front end carrier gas line.
The system usually consisting of an on-off toggle valve and a restriction that reduces the flow to 1-2 ml/min that is plumbed across the usual carrier on-off valve. You open the bypass, and close the normal valve. It can all be combined in a multi-port valve, if the savings are justified.
Why not go the whole hog, and suggest a $2,000+ time-controlled, auto-switching, multiport valve to save $100 of gas per year?. It should be worth some sort of performance bonus
I'd also endorse using the most appropriate grade of carrier gas for TCD and FID. I used welding grade argon and technical grade hydrogen and helium on packed and capillary column GCs with gas purifiers that were regenerated once a year. In general, most carrier gas disappears through leaks or high split vents when instruments aren't running, and using cheap gases meant instruments were always ready to go.
You need ultrapure gases if you are looking for trace impurities or have oxygen-sensitive columns or detectors. If you are buying ultra-pure gases, invest in a leak detector and some simple bypass valve setups.
Bruce Hamilton
-
- Posts: 3594
- Joined: Mon Aug 30, 2004 11:43 pm
Like chromatographer1 stated, I'd also just lower the helium flow. Keep it simple.
-
- Posts: 543
- Joined: Thu May 18, 2006 2:36 pm
Another way of looking at it is (somebody - please check my calculations!)
A standard 1.5m (5ft) cylinder of helium holds 9 cu.m
=9000000mls
at 30ml/min =300000mins=5000hrs=200 days approx. (and assuming that nobody in the lab is talking in Donald Duck)
Hope that helps
Ralph
A standard 1.5m (5ft) cylinder of helium holds 9 cu.m
=9000000mls
at 30ml/min =300000mins=5000hrs=200 days approx. (and assuming that nobody in the lab is talking in Donald Duck)
Hope that helps
Ralph
-
- Posts: 1233
- Joined: Tue Oct 11, 2005 7:20 am
A+ for mathsAnother way of looking at it is (somebody - please check my calculations!)
A standard 1.5m (5ft) cylinder of helium holds 9 cu.m
=9000000mls at 30ml/min =300000mins=5000hrs=200 days approx.
One "advantage" of using a bypass system, for both capillary and packed column GCs, is that instrument settings are not changed, consequently retention times are almost identical day to day.
That "advantage" may no longer exist with modern electronic pressure or flow controls, but is very valuable for older and simpler GCs ( eg HP5890 ).
Please keep having fun,
Bruce Hamilton
-
- Posts: 543
- Joined: Thu May 18, 2006 2:36 pm
Thanks Bruce. Recently, when I expressed concern about the amount of helium that we were losing through leaks in the piping (due to the numerous couplings used in the supply to the lab from a remote location) my boss replied "Can't we use a helium generator?" !!! Imagine the sunburn!
Regards,
Ralph
Regards,
Ralph
-
- Posts: 3594
- Joined: Mon Aug 30, 2004 11:43 pm
That boss sounds like Dilbert's boss (US comic strip about the work world).
-
- Posts: 122
- Joined: Fri Sep 17, 2004 1:17 pm
I liked this one, it's posted in our office:
Dilbert's boss - What's this MFU2 on your timeline?
Dilbert - That's management foul up number 2 - it usually happens around the third week.
Dilbert's boss - We don't anticipate any management mistakes.
Dilbert - That's MFU1.
Dilbert's boss - What's this MFU2 on your timeline?
Dilbert - That's management foul up number 2 - it usually happens around the third week.
Dilbert's boss - We don't anticipate any management mistakes.
Dilbert - That's MFU1.
11 posts
Page 1 of 1
Who is online
In total there are 18 users online :: 1 registered, 0 hidden and 17 guests (based on users active over the past 5 minutes)
Most users ever online was 5108 on Wed Nov 05, 2025 8:51 pm
Users browsing this forum: Baidu [Spider] and 17 guests
Most users ever online was 5108 on Wed Nov 05, 2025 8:51 pm
Users browsing this forum: Baidu [Spider] and 17 guests
Latest Blog Posts from Separation Science
Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.
Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.
- Follow us on Twitter: @Sep_Science
- Follow us on Linkedin: Separation Science