if calibration curve is not linear, can I still use it?

Discussions about GC and other "gas phase" separation techniques.

4 posts Page 1 of 1
I noticed people were talking about quadratic calibration curves on non-polar column. I had the same problem. I tried to play with inlet and interface temperature and even ramp temperature, but was not able to get a linear calibration curve. If we can not conveniently change to another column, can I use the quadratic curve to quantitate my real sample?

thanks a lot,
ziman

By definition calibration curve :)
Ideally you should have a straight line however it is not unusual to get non linearity at the ends of the line. If the calibration is actually a curve then you should make sure you have sufficient points that you are confident that any compounds you quantify are accurately defined by the calibration curve. With a straight line you can get away with just three calibration points :roll: with a true curve you will need a lot more.

You can use any shape of calibration that you want, as long as you have enough points - the sharper the curve the more points you need.

BUT - can you be sure that the mass - signal relationship that you get from the standards that you use to construct the curve will be the same for the analyte in a real sample. In this particular case the assumption that they will be the same might not be a robust one because there is a good chance that the adsorption of analyte that is causing the deviation from linearity with the standards will not be same when you run samples.

You need to spike samples with analyte and determine recovery (using the quadratic curve calibration) at high, medium and low concentrations. If it stays the same you are OK.

Peter
Peter Apps

thank you for your reply.
My curve is not non linear at the end, but at the beginning, then when the concentration is bigger, it is more linear. slope of the tangent is increasing instead of decreasing according to increasing concentration. Tt is flat at low concentration and steep at high concentration. Is that due to the active adsorption spots in my non-polar column?

thanks a lot for your time
ziman
4 posts Page 1 of 1

Who is online

In total there are 13 users online :: 0 registered, 0 hidden and 13 guests (based on users active over the past 5 minutes)
Most users ever online was 5108 on Wed Nov 05, 2025 8:51 pm

Users browsing this forum: No registered users and 13 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry