Thanks for posting the chromatograms, very helpful.
Because your detector is only seeing some of the compounds ( if you split the column outlet and send to FID as well, you will see the organics, but not any water ), quenching may occur because of coeluting peaks.
Your standards should be processing the same as your samples, and in the same solvents. Linearity should be good for both peaks of interest, and you don't have to worry about other peaks, besides ensuring they elute in the same run

.
You need to ensure that reducing the resolution will not cause coelution. The detector should not become contaminated - provided it is at a reasonable temperature ( usually recommended by instrument manufacturers ).
I'd usually select a thin film column with low bleed for ECD of a few halogenated molecules and, in my experience 624 columns tend to bleed more than many other phases. Low bleed = low background, less problems and detector contamination, and usually means you can inject less.
I'm not sure where you method came from, but if it works please keep using it.
Bruce Hamilton