MP: Time(min) A (KH2PO4 25mM,ph 2.
0 2% 0 98%
3 2% 3% 95%
3.10 2% 20% 78%
15 2% 20% 78%
20 2% 0 98%
30 2% 0 98%
Detector:UV(260nm)
Flow rate: 0.8ml/min
Oven Temp: 25℃
Column: Zorbax SB C18 250X4.6mm, AQ
In the test for the first time, the chromatography peak was normal,but in the next day, the sampler was injected, the chromatography peak was very broad, the condition of chromatography is absolutely same to the previous test, even the sampler and MP were not changed either.
In the last, the equilibrium time was extened 20 minitues, in other word, the 30min in last line of time table was modified as 50 min, the peak was normal.
But the peak was very broad again this day, even if the euilibrium time was extened 40 min, the peak was till very broad.
Anyone can tell me what reason it is, how should i solve the problem?
Thanks in advance!