Tim » Thu Oct 28, 2004 3:37 pm
We have to use Multidraw loops for a few of our methods - mainly for low level analytes, e.g. in cleaning validation samples. For this they work fine - it does what it suggests, draws up multiple quantites of sample and ejects them into the larger seat capilary (400 or 1400ul, giving a total possible volume of 500 or 1500ul).
The only issues are that it takes quite a bit longer to perform each injection, since the 1100 has to pull up 100ul, move the vial, eject the volume into the capillary, move the vial back, draw up the next volume, move the vial, etc. This only really becomes an issue where you have a really short run time, since this would increase you injection-injection time.
The other issue, which must be considered is that you need to remove the seat capillary again after you have completed your large volume injections, unless you are going to dedicate a system to just doing large injections. If you don't, you get a large dead volume which causes increased extra-column band broadening and can lead to failing system stuitability parameters (N and T typically).
I don't have any experience of the large loop/syringe option, but guess it would be faster for each injection, possibly give similar problems with smaller injection volumes and also there is a risk of decreased accuracy/precision on low volume injections, since the syringe would only be pulling up a small volume, so would be subject to greater error from the stepper motor/gearing, etc.