Dear people,


I am using a GC-TCD with a CTR-III packed column to analyze Ar, N2, O2 in air.
We used to obtain the O2 content substracting the Ar peak area to the Ar+O2 peak area. Recently I realized that the results of the CTR-III column were not allowing me to quantify the O2 because the Ar+O2 peak was being smaller than the Ar only peak area.

I thought the problem could be solved conditioning the CTR-III column. The tech papers I found say that conditioning must be made increasing the carrier flowrate to 40mL/min and heating the column to 140C for one hour. Then increasing the temperature at a rate of 5C/min to 250C and holding it there for 12hours.
The problem comes from the label that the CTR-III column has, where it is said that the MAOT is 200C when every data sheet says the max temp is 275C.

Please, could anybody explain me which one is the correct max temperature for the CTR-III packed column??

Thanks a lot,
Zango:)

Hi Zango,

The literature does indeed suggest 275°C as the MAOT. I would contact Alltech and and ask them.

Regards,

Ralph
IMG]http://i5.tinypic.com/24chzc0.jpg[/IMG]

http://i5.tinypic.com/24chzc0.jpg

Sorry for the commercial but in case anyone else is interested in separating these three gases at room temperature here is a sample chromatogram using a packed column at 28°C.

Rodney George
Senior Research and Development Scientist
Gas Separations Research
Supelco
595 North Harrison Road
Bellefonte, PA 16823

814-359-5737 voice
814-359-5459 fax
rgeorge@sial.com
http://i5.tinypic.com/24ci72v.jpg

This link is to the chromatogram of air. The previous one is Neon, Argon, Oxygen, Nitrogen, and Krypton.

The column is a 9m 1/8" packed column with special packing on a custom basis using a µTCD 6890 GC.

Rodney George
Senior Research and Development Scientist
Gas Separations Research
Supelco
595 North Harrison Road
Bellefonte, PA 16823

814-359-5737 voice
814-359-5459 fax
rgeorge@sial.com
4 posts Page 1 of 1

Who is online

In total there are 18 users online :: 1 registered, 0 hidden and 17 guests (based on users active over the past 5 minutes)
Most users ever online was 5108 on Wed Nov 05, 2025 8:51 pm

Users browsing this forum: Baidu [Spider] and 17 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry