GC 7980A INLET PRESSURE SHUT DOWN
Discussions about GC and other "gas phase" separation techniques.
4 posts
Page 1 of 1
Hi, the gc 7980a fid system I am using seems to take longer time to attain the inlet pressure and then the system down.All rthe white the system beeps with error message low inlet pressure.Once the system is restarted again the pressure is attained at once .but this seems to happens everyday while switching on the system.can anyone help me out with this?
-
- Posts: 3477
- Joined: Mon Jan 07, 2013 8:54 pm
Does the inlet cool to room temperature each night when shutdown?
If so it could be that on the second start the inlet has had time to heat up and is sealing a leak that happens when cold. The expansion of the metal or seals is sealing the leak when it is hot.
With the system on, turn off the inlet heater and allow the inlet to cool to room temperature and see if the pressure holds. If not, then check for leaks and fix those and the problem should go away.
If so it could be that on the second start the inlet has had time to heat up and is sealing a leak that happens when cold. The expansion of the metal or seals is sealing the leak when it is hot.
With the system on, turn off the inlet heater and allow the inlet to cool to room temperature and see if the pressure holds. If not, then check for leaks and fix those and the problem should go away.
The past is there to guide us into the future, not to dwell in.
-
- Posts: 517
- Joined: Thu Jul 14, 2016 12:08 pm
There is no reason to shut down a GC overnight if you use it everyday. You want to maintain a (low) flow of carrier gas 24/7 so your column is not collecting air/water/dust.but this seems to happens everyday while switching on the system.can anyone help me out with this?
Several causes for the GC not being able to maintain a stable pressure:
- A blockage
- A leak
- A pressure that is outside the range that system can provide
- Method settings that are completely off: the system calculates the pressure needed to provide a flow that the user wants based on the dimensions of the column provided by the user.
-
- Posts: 326
- Joined: Wed Oct 02, 2013 12:20 pm
Hello
If you have gauge regulated gas supply (cylinders) for carrier gas it shouldn't happen.
Are you using gas generator for carrier gas? (Hydrogen)
If you turn it on with GC (inlet is on) generator needs time to produce gas - if you have generator it could be reason for inlet shutdown.
Regards
Tomasz Kubowicz
If you have gauge regulated gas supply (cylinders) for carrier gas it shouldn't happen.
Are you using gas generator for carrier gas? (Hydrogen)
If you turn it on with GC (inlet is on) generator needs time to produce gas - if you have generator it could be reason for inlet shutdown.
Regards
Tomasz Kubowicz
4 posts
Page 1 of 1
Who is online
In total there are 21 users online :: 1 registered, 0 hidden and 20 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: Google [Bot] and 20 guests
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: Google [Bot] and 20 guests
Latest Blog Posts from Separation Science
Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.
Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.
- Follow us on Twitter: @Sep_Science
- Follow us on Linkedin: Separation Science