WK,
From my experience working with organic acid analysis, the column needs to be compatible with pure or highly aqueous mobile phase and hydrolytically stable at operating pH, say pH2. The columns that are potentially useful for you include Atlantis dC18 (Waters), Stablebond C18 (Agilent), and Acclaim OA (Dionex). I am not sure if Lune(2) C18 is compatible with 100% aqueous.
Once the column (RP) is chosen, I would consider two factors in method development stage - mobile phase pH and ionic strength. I have much experience with Acclaim OA column for organic acid analysis. Here is the chromatographic condition I used to separate oxalic, tartaric, malic, formic, acetic, lactic, citric, succinic, and fumatic acids in a single run.
Column: Acclaim OA, 5um 4x150 mm
Mobile phase: 50 mM phosphate buffer, pH2.7
Flow rate: 0.6 mL/min
Temperature: 30 C
Detection: UV at 210 nm
If you need to increase the resolution between acetic and lactic, raise pH up to 2.9. Other more hydrophobic analytes can be eluted with ACN gradient.
By above method, I haven't observed any retention time reproducibility problem so far except one time I did something wrong in preparing my mobile phase.
I hope my reply is helpful, and am willing to share more information with you.
Good luck,
XL