I'm curious how to best inject methane to measure column flow. I've seen a setup that Restek sells for filling a manual syringe with methane, but on my 5890 I wouldn't trust my injection technique to get a good retention time. Is there a clever way to do this with a autosampler and a vial?
-
- Posts: 960
- Joined: Thu Apr 18, 2013 2:10 pm
You can trust your injection technique. For setting the flow, you don't care about how well you can reproduce the size of the methane peak, you just care about how fast it comes out. As long as your detector can see the methane peak, that's really all you need.
-
- Posts: 5433
- Joined: Thu Oct 13, 2005 2:29 pm
In this case clever = simple. Blow some methane into the vial, put the septum cap on, inject. As long as you see an on-scale peak you can get the dead time.I'm curious how to best inject methane to measure column flow. I've seen a setup that Restek sells for filling a manual syringe with methane, but on my 5890 I wouldn't trust my injection technique to get a good retention time. Is there a clever way to do this with a autosampler and a vial?
Peter
Peter Apps
-
- Posts: 50
- Joined: Sun Nov 24, 2013 6:27 am
Blowing methane in the vial was my original plan, but it's inevitable that oxygen will get in the vial, and I'm trying to not to introduce oxygen into the column. On the other hand, even if you load the syringe with pure methane, it is certain that some large percentage will exchange with air before the injection occurs.
Is oxygen really not that harmful in these amounts? I've heard of simply injecting air to set column flow.
Is oxygen really not that harmful in these amounts? I've heard of simply injecting air to set column flow.
-
- Posts: 5433
- Joined: Thu Oct 13, 2005 2:29 pm
Under reasonable conditions (1ul injection, 10:1 split, 2 ml/min flow) the oxygen concentration in the carrier gas is 10 micromoles/mole and the column is exposed for 2 min (or a higher concentration in the air plug but for the peak width in time). I doubt it would be a problem.
Peter
Peter
Peter Apps
-
- Posts: 543
- Joined: Thu May 18, 2006 2:36 pm
Hi
Try using a standard disposable cigarette lighter (flint based type) as a supply of your non retained gas without introducing oxygen
The nozzle bore was designed to accept a standard 10uL syringe needle
Regards
Ralph
Try using a standard disposable cigarette lighter (flint based type) as a supply of your non retained gas without introducing oxygen
The nozzle bore was designed to accept a standard 10uL syringe needle
Regards
Ralph
Regards
Ralph
Ralph
-
- Posts: 50
- Joined: Sun Nov 24, 2013 6:27 am
What a great idea! I will give that a try.
7 posts
Page 1 of 1
Who is online
In total there are 30 users online :: 2 registered, 0 hidden and 28 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: Ahrefs [Bot], Semrush [Bot] and 28 guests
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: Ahrefs [Bot], Semrush [Bot] and 28 guests
Latest Blog Posts from Separation Science
Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.
Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.
- Follow us on Twitter: @Sep_Science
- Follow us on Linkedin: Separation Science