I am seeing a 10 fold increase in response when I use an unpacked single tapered liner instead of my usual glass wool packed single tapered liner (SGE Focus Liner). I expected to see some difference but not as much as this. Anyone got an explanation.

Cheers

Martin,

Would you add what the analyte is and its concentration.

Thanks.

oh - and is the solvent peak staying the same width?

Ralph

split or splitless ?
Peter Apps

I am using an in-house test mix containing compounds with a variety of functional groups at concentrations of approx. 0.1% in acetone. Injections are 100:1 split. All compounds are affected equally.

Hi Martin

Like a lot of people posting problems you are making this difficult by supplying information one little piece at a time as if you are a suspect being interrogated !!

Please supply FULL details of your set-up and methods, i.e. column dimensions, flow rates, injection manual or auto etc etc so that a diagnosis does not have to be based on guesswork.

Thanks Peter
Peter Apps
6 posts Page 1 of 1

Who is online

In total there are 25 users online :: 1 registered, 0 hidden and 24 guests (based on users active over the past 5 minutes)
Most users ever online was 5108 on Wed Nov 05, 2025 8:51 pm

Users browsing this forum: Bing [Bot] and 24 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry