It is my supposition that you injected a large amount of pure hydrogen onto your 13X column.
The tailing you are seeing is the hydrogen slowly eluting from the pores of the MS 13X.
When you try to measure hydrogen with a carrier gas that is close in thermal conductivity certain concentration effects can occur. This is known historically as the "w" effect.
This effect is published in the literature when using helium as a carrier. I suspect you went beyond the limits of hydrogen/argon mixture linearity when you injected your hydrogen sample, especially since it appears that you are not seeing a return to baseline.
While the use of Argon allows you to have a measurement of hydrogen and other gases concurrently, the dynamic linear range may not be adequate for your needs.
Another reason for the tailing is an amount of water or some other contaminant in your system or hydrogen gas sample which may be eluting slowly from the column.
This tailing diagnosis assumes that no defect is present in your packing material or in the packed column made from this packing material.