Unless you have to use gradient, isocratic is alway preferred, often because you can use a RI detector, and you can also use recycle facility with relative impunity. Desirable if you plan to recover and reuse solvent as well.
However, the critical aspect is that you have to resolve the impurity from other impurities, and you have to elute ( or backflush ) all the other material before the next injection. If gradient is required then use it,. Gradient also may reduce solvent use, and that's preferred.
Firstly, you have to determine optimum resolution and retention, then column size options become apparent.
The typical maximum column size is 50mm diameter, which usually is getting up towards 100 ml/min. If you want to purify large amounts, and you have plenty of resolution, consider using a Biotage column system, they go up to 400mm diameter, but don't usually have the resolution for closely-eluting compounds. Cheaper and faster than larger HPLCs, but absolutely no use if you can't get the separation.
In my limited experience, fewer larger runs are preferable to many smaller runs, especially if you've optimised the method for rapid elution and fast cycle time. If you are going to submit sample data to regulators, you may need to show that all runs and feeds were similar.
I'd also caution about response factors - yes, they can be used as a guide, but remember that balance uncertainy bites deep when weighing a mg or two on a 5 figure balance, and you still have to assay the impurity to derive a purity for the solid to generate the factor.
The choice of using synthetic chemists depends on your company, if it's an option, and the impurity is needed for future work, that would be an option I'd favour. Nothing beats heaps of crystals

.