GC/ECD , PEACKS
Dear all friends,
May you please help me. I have GC/ECD and I make an empty run but I see periodical peacks exactly like electrocardiography (heart works' print).
I made my detector conditioned (oven:40 0C when detector 400 0C), but nothing changed.
When I do an empty run instead of a flat and horisontal line but it is one peack after each 2 minutes. The peacks height go about 70-80 Hz.
What can I do.Do you know why it works like this?
Thank you for your help.
Semiramis
I have no idea what is causing your problem however I can ask you to supply more details and try a simple test. What GC, ECD and detector conditions are you using ? We need to find out if the peaks are coming from the injector/column or the detector. Disconnect the column from the base of the ECD and cap it off. Does the output still have a peak every two minutes ? If yes then problem is most likely either pulsing in the detector gas or fault in the detector. If no the problem is either a real peak coming through or again a gas pulse/contamination.
Regards Richard
Dear Richard,
Many thanks for your kind reply.
My GC is Chrompack, CP9001.
I have done detector conditioning but again problem exist.
Ok.I try to use your recommendation and see if the problem removes or not.
Thank you.
Semiramis
In total there are 23 users online :: 2 registered, 0 hidden and 21 guests (based on users active over the past 5 minutes)
Users browsing this forum: Ahrefs [Bot], Bing [Bot] and 21 guests