Hi! I'm running an application with a mobile phase consisting of 95% of the following buffer: pyrophosphate sodium decahydrate 10mM, Na2EDTA 10mM, Triethylamine 0.1% and 5% ACN. The column is an ODS C18. The problem is that the maximum column life is 1 week. What do you advice me to do? The problem here is that with this mobile phase I have the best chromatography for my analyte.
Thanks
Magda[/quote]
Magda

Hello Magda,

It may be the best conditions for your separation, but the worst for your column!
Silica dissolves really easy at this pH.

There a few opinions, but the two most realistic are:
• In case you need separation efficiency, switch to hybrid type of columns (e.g. Gemini/XBridge)
• In case column efficiency is not important, switch to polymeric materials (e.g. PLRP-S).

If you want to stick to this aggressive mobile phase, you have to change the column !!
2 posts Page 1 of 1

Who is online

In total there are 15 users online :: 2 registered, 0 hidden and 13 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Baidu [Spider], Semrush [Bot] and 13 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry