MS related peak shape
Discussions about GC-MS, LC-MS, LC-FTIR, and other "coupled" analytical techniques.
3 posts
Page 1 of 1
I am operating a GC-MS in NCI mode for analysis of PBDEs. Major peak tailing was observed after cleaning the source (no other maintenance performed). The source is normally operated at 150C; at source temp 250 and 300, the peak shape dramatically improves. At source 150, the peak shape for the molecular ion is great, while monitoring sequential mass fragments in the same injection, the peak shape will deteriorate. This was sudden and cannot be explained. A second, new source was installed with no improvement. An identical instrument in the same lab has no problems operating at 150 under these conditions.
-
- Posts: 517
- Joined: Thu Jul 14, 2016 12:08 pm
Not sure what your question is
It seems that either
- You baked out your source by going to 250-300. It's not a bad idea, especially after cleaning, to approach the upper temperatures of your source for at least a couple of hours.
- The ionization of these molecules is just better at higher temperatures.
I don't work with NCI, but for instance for PAHs with GC-EI-MS, i notice a significant improvement in peak shape for the heavy ones going from 200 °C source temperature to 280 °C.
It seems that either
- You baked out your source by going to 250-300. It's not a bad idea, especially after cleaning, to approach the upper temperatures of your source for at least a couple of hours.
- The ionization of these molecules is just better at higher temperatures.
I don't work with NCI, but for instance for PAHs with GC-EI-MS, i notice a significant improvement in peak shape for the heavy ones going from 200 °C source temperature to 280 °C.
-
- Posts: 5433
- Joined: Thu Oct 13, 2005 2:29 pm
Welcome to the forum.
Are you asking about chromatographic peaks, or mass spectrum peaks ?
Peter
Are you asking about chromatographic peaks, or mass spectrum peaks ?
Peter
Peter Apps
3 posts
Page 1 of 1
Who is online
In total there are 23 users online :: 0 registered, 0 hidden and 23 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: No registered users and 23 guests
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: No registered users and 23 guests
Latest Blog Posts from Separation Science
Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.
Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.
- Follow us on Twitter: @Sep_Science
- Follow us on Linkedin: Separation Science