Wet Sample Chamber- Alliance

Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.

4 posts Page 1 of 1
Is anyone aware of the cause of this problem? Initially we noticed the tops of vials being slightly wet, but now the entire sample chamber is wet with mobile phase.

Thanks!

peter

mobile phase? Are you sure that it's mobile phase?
It could be condense when cooling the chamber. After a while there is some ice forming, and then you can have much condense.

If it's mobile phase it could be that there is a tubing disconnected inside the HPLC. You can easily see when screwing off the right side of the Alliance.

good luck

Bart

Not entirely sure it's mobile phase. It could be condensation, but I would expect it on our other machines which are identical.

It could be a mobil phase or a needle wash solution. If your eluent contains buffer check the pH.
Prime the system and run purge with compressibility check . If the test is pass the needle seal is leaked, if not the injector seals are leaked. The needle wash seal can change easily. The injector seal can be changed by service engineer.
4 posts Page 1 of 1

Who is online

In total there are 15 users online :: 2 registered, 0 hidden and 13 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Baidu [Spider], Semrush [Bot] and 13 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry