We're running a method on an Acquity H-Class system that is generally isocratic (at least for our peaks of interest) with 0.1%TFA in 91:9 Water:Acetonitrile. When we pre-mix the mobile phase so that the method just runs at 100% A, the baseline looks amazing. However, when we prep MPA as 0.1% TFA in Water and MPB as 0.1% TFA in Acetonitrile and run it as 91% A and 9%B, the baseline is unacceptable for what we need it for (low-level impurity detection/quantitation). We've run both (pre-mixed and instrument-mixed) on several Acquity systems in our lab and have seen the same results.
Is our 100uL mixer inadequate? Does anyone have any suggestions on improving the baseline and keeping the simplified MBA/MPB approach? The flow rate is set at 0.6mL/min - 225nm detection wavelength. See below for a screenshot of baselines (blank injections).