I need your help! Any suggestions are welcome, please!
I am working on the quantitation of an oligosaccharide in plasma. The analyte has 24 carbon and MW around 600 Da. The instrument is LC/MS/MS. This compound is too hydrophilic to retain by the RP-HPLC. I use Luna amino column (150x2.1). I start with 75% B (ACN), hold for 0.3 mins and drop to 30% ACN for 2.4 mins and back to 75% B to 6.5 mins. Right now the retention time is 4.3 mins. I have two major problems right now:
1. The reproducibility is bad, even in solvent standard. For solvent standard of 20 ng/mL, the peak height was shift from 270 to 190 during three continue injections. However, at higher concentration like 1000 ng/mL, the variation is smaller. The RT has no significant shift (maybe 0.02 mins).
Here are some tests I performed:
I. Extent each run form 6.5 to 8.0. No luck
II. Change the gradient like 90-30, 85-30 and 75-30. Same problem.
III. Change the mobile phase, A from 0.1% FA in water to 5 mM NH4OAC (pH=5.0). Same problem.
IV. I also try CN column. Similar problem was noticed.
One concern is my Massspec is not stable. But it is fine when I use same mobile phase to analyze other compound. I still feel that this is a LC problem. Whenever the sensitivity was change, the peak shape was change too. The lower sensitivity, the more significant tailing was observed.
I wonder if I can try something else. Shall I try ion pair reagent? Which ion pair reagent you can recommend for the LCMS system?
2. Significant matrix effect was observed. After inject 10 or more matrix samples, the signal of solvent was significantly dropped (about 70%). Even after long time washing and inject solvent standard, the sensitivity recovered slowly.
Any suggestions?