I have had success retaining glycolic acid on a polar endcapped C18 column. It sounds like you get decent retention of the hydroquinone with or with out ion pair reagents. I used the Phenomenex Synergi Hydro RP. Should work for both compounds with a pretty simple MS compatable mobile phase.
-
- Posts: 480
- Joined: Fri Sep 03, 2004 7:57 am
HILIC does not work, both peak could not be separated though we used 95% acetonitrile or methanol in mobile phase.
I will back to RP-LC, using gradient run from 100% aqueous mobile phase.
I will back to RP-LC, using gradient run from 100% aqueous mobile phase.
-
- Posts: 2846
- Joined: Mon Aug 30, 2004 7:17 am
Could it be that now you gave up too early on HILIC regarding variation of pH? Doesn´t make sense that your compounds should not separate.
-
- Posts: 156
- Joined: Tue Sep 07, 2004 8:23 pm
In what solvent are your samples?
Try to dissolve in as much acetonitrile as possible or reduce injection volume for water samples.
In HILIC a large volume injection of sample in water is just as bad as a large volume injection of sample in acetonitrile on a RP column.
Try to dissolve in as much acetonitrile as possible or reduce injection volume for water samples.
In HILIC a large volume injection of sample in water is just as bad as a large volume injection of sample in acetonitrile on a RP column.
------------------------
Merck SeQuant AB
http://www.sequant.com
Merck SeQuant AB
http://www.sequant.com
-
- Posts: 2916
- Joined: Mon Aug 30, 2004 10:19 pm
syx,
I would use a quick gradient from 90% acetonitrile to 50% acetonitrile on your silica column. Once you know where your compounds elute, you can use an isocratic run to optimize everything. I do not recommend to use any acid in your first runs, since the acid in the mobile phase will pull the analytes into the mobile phase and reduce retention. If there is need to do so, you can always add some acid later during finetuning.
I would use a quick gradient from 90% acetonitrile to 50% acetonitrile on your silica column. Once you know where your compounds elute, you can use an isocratic run to optimize everything. I do not recommend to use any acid in your first runs, since the acid in the mobile phase will pull the analytes into the mobile phase and reduce retention. If there is need to do so, you can always add some acid later during finetuning.
-
- Posts: 480
- Joined: Fri Sep 03, 2004 7:57 am
Wow, thanks for the responds.
We have used mobile phase as the solvent. In last experiment we did not use acid in the mobile phase, just a mixture of water and organic solvent. Both analytes were eluted too early, about 3.7 minutes.
We have used mobile phase as the solvent. In last experiment we did not use acid in the mobile phase, just a mixture of water and organic solvent. Both analytes were eluted too early, about 3.7 minutes.
21 posts
Page 2 of 2
Who is online
In total there are 17 users online :: 2 registered, 0 hidden and 15 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: Google [Bot], John Guajardo and 15 guests
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: Google [Bot], John Guajardo and 15 guests
Latest Blog Posts from Separation Science
Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.
Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.
- Follow us on Twitter: @Sep_Science
- Follow us on Linkedin: Separation Science