I am conducting studies on drug elution from various formulation matrices. All my samples pass through a C18 SPE procedure prior to injection in HPLC. Over time the back pressure in the HPLC system increases. I'll take a guess that I see a rise from 1400psi to 2000psi in 200 injections. Replacing the guard column lowers the pressure, but the pressure builds back up quickly. The retention time does not seem to move.
The only thing that I can think that is coming from my samples that may cause problems are oils or lipids. The lipid must be molecularized in order to pass through the SPE I guess. I have seen the lipids solidify in the HPLC sample vial when cooled though.
HPLC method 1: MeCN:water 60:40, flow 1ml/min, C18 column
My questions are:
1. could this pressure rise be seen as "normal"?
2. is it likely that an oil such as castor oil, or a lipid cause backpressure increase?
3. is it common for labs to implement a regular column cleaning procedure? for example, run 95% acetone, IPA, etc for 30min after each shift/sequence?
thanks