Well, it looks like it is a contaminant from the pH probe used during pH adjustment of the mobile phase. I am guessing the preservative in the VWR buffers (Dowicide A) is the culprit (not to mention the pthalate in the pH 4 buffer). The main Amax's we were seeing in our mystery peak were at 245 nm and 285 nm which match up nicely with the spectra of the pH 2 and 7 standardization buffers.
We are now looking into preservative free pH solutions as well as different types of probes that we can clean better prior imersion in the liquid.
I just wanted to give you all the heads up if you too were chasing after a mystery peak.