I'm having a strange issue with one of my Purge and trap systems. I am running EPA 8260 method on a 6890/5973 with an OI 4660 concentrator and Archon 8100 auto sampler. What has been happening is a loss of about 75% intensity on the first 2/3 of the chromatogram. Once we hit the heavier substituted benzenes, starting with ethylbenzene and going through 1,3,5-trimethyl benzene, the recoveries return to normal. All the internal standards follow the same trend. I've exhausted my knowledge on what could be causing this.
so far I have changed the trap, we use #10, changed the Sulfinert transfer lines the water management unit sonicated and baked the 6-port valve, sparge mount and 4 way cross filter. I have also replaced the pencil filter and purge needle for the auto sampler and finally I swapped the Volatiles Interface for the GC inlet.
I spent a few runs watching the pressure, flow and temperature readings on all parts of the system and everything is normal. I've also tried using both an electronic leak detector on external pieces and a duster can on the MSD seals and found nothing leaking.
An interesting note is that a few days ago the first calibration check, which we run at 25 ppb, would recover great and all subsequent runs were back down to 25% in the first 2/3 of the run and now even the first after a nights rest is showing the symptoms.
At this point I'm out of ideas. any insight would be hugely appreciated.