My problems are two fold. (at least)
1st, The retention time of my analyte (quaternary amine) is drifting inward. It appears to be related to the column lot. I have contacted Dionex but have not gotten a response. The two new columns that I have tried start out with about 15 minute retention time and then drift in about 0.1 min per injection. This continues until all retention is lost. I have a second method for the same compound. It uses an eluent of 15 mM MSA, 1% acetonitrile and 10% methanol. This produces stable retention for the same compound. I have increased the acid concentation and observed improved peak shape but still have the retention time variation issue.
My second problem comes from trying to find a solution to the first.
I experience a large increase in noise (up to 20 time larger) if I change the flow rate of the eluent. The method flow rate is 1.5 mL/minute. I have tried flow rates from 1.2 mL/minute to 1.75 mL/minute. The noise does not appear to be influenced by the flow rate of my regenerant.
Thanks in advance for all responses.