Hi,

I have got a new toy in my lab, namely ELS-detector (evaporating light scattering). I have found that the precision and linearity are not in the same level than with a UV-detector. For example the RSD for six consecutive injections of pure standard is about 6.5 % and showing an increasing trend (carry-over is not the cause since the blank is ok). Also an increasing trend of response factors (peak area / mass) is observed with the linearity test.

Has anybody noticed same kind of problems / features with ELS-detectors?

Thanks in advance!

Sami

Is it a "regular" ELSD or a Corona Discharge unit?

Are you running isocratically? Are your RTs consistent?

We need more details.
Thanks,
DR
Image

I suppose it is regular detector (Sedere Sedex 85). I have a gradient run, using Waters HILIC column with 10 mM NH4COOH (eluent A) and ACN (eluent B). Retention times are ok, RSD for those is about 0.3 %. The peak characteristics (form, tailing factor, plates) are all fine, so I think the reason is on the detector, not in the chromatography process.

Sami

Area count precision is always worse on an ELSD. However, 6.5 % is on the high side. I would expect (assuming the autosampler and pump are working correctly) area counts of about 3% RSD. ELSD's are NOT linear (you can not change the laws of physics), fit the data to a quadratic equation. To troubleshoot the issue, put a UV detector in-line before the ELSD and inject a compound that will give response in both. I would expect <0.5% area from the UV and <3.0 from the ELSD.
4 posts Page 1 of 1

Who is online

In total there are 15 users online :: 2 registered, 0 hidden and 13 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Baidu [Spider], Semrush [Bot] and 13 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry