Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.

2 posts Page 1 of 1
Gradient system from 0% to 100% B over 95 minutes. Low organinc to high organic Mobile phase progression.

Data was collected for 110 min. The gradient baseline drift is not excessive. The dwell time is about 3 min so at 98 min the baseline comes down.

What would cause a nice sharp peak to appear at 107 min. It is seen in the standard and sample but not the blank.

The analyst in charge of the system is wondering if its a late eluting impurity. My response was a 99% no, not possible.

Why would a late eluting (non-polar) peak that has retained that long wait until the conditions are as they were in the beggining of the run to elute.

Is the peak there if you increase the hold time at the end of the gradient?
Thanks,
DR
Image
2 posts Page 1 of 1

Who is online

In total there are 15 users online :: 2 registered, 0 hidden and 13 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Baidu [Spider], Semrush [Bot] and 13 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry