There are reversed-phase columns that can handle the large proteins. You need 300-1000A pore size, either silica or polymer, or a monolith. The problem is that RP elution conditions denature the proteins, and that creates other issues. Most people in the business use ion exchange for antibodies. (Dionex ProPac is an excellent choice.) Hydrophobic interaction chromatography is the other good choice.
Mark,
Thanks a lot for your reply. I am using Dionex ProPac for the purpose now. The problem I have with this CEX method is that the separation of different form of the antibody is not even close to baseline separation. One possible reason might be the purity of the sample was not good. And also there are a bunch of peaks clattered within 10 minutes in a 60 minute run. There was always a long tail. That's why I am looking for something else to try.
In RP-HPLC, the protein is denatured. But I am not interested in recovering the protein anyway. As long as it detects or separates the change, it works for my purpose.
BTW, I attended an interesting seminar where the presenter said they could use a large column to purify protein with RP-HPLC (protein recovered without denatured). He did not reveal details.