Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.
21 posts
Page 2 of 2
I have also had experience with a couple of episodes of bacteria growing where it should be impossible. One instance was due to a technician only hearing the clean portion of clean and sterilize before using the glassware instructions. Since bacteria grow in the solvents, filtering through a membrane filter can be a simple and useful precaution.
-
- Posts: 2175
- Joined: Tue Aug 17, 2004 7:59 pm
High background absorbance - but that has to come from a larger colony, and there's no room for that *in* a system, only in the jugs.
Thanks,
DR

DR

-
- Posts: 23
- Joined: Thu Aug 25, 2005 5:50 am
How about a bacterial colony in the column?
The reason I ask this is that I have seen the same very broad peak (max at 205nm) eluting from all of my RP columns, even after thorough flushing (and storing in 70 - 100% ACN).
I have always assumed that it is an impurity from my sample, maybe a very strongly retained protein that I have not managed to flush out..
I have not managed to get rid of it completely. It's still stubbornly lurking in my columns after every analysis. So far, I've only been successful at reducing the peak area (after prolonged cleaning with H2O/ACN gradients).
Any ideas?
The reason I ask this is that I have seen the same very broad peak (max at 205nm) eluting from all of my RP columns, even after thorough flushing (and storing in 70 - 100% ACN).
I have always assumed that it is an impurity from my sample, maybe a very strongly retained protein that I have not managed to flush out..
I have not managed to get rid of it completely. It's still stubbornly lurking in my columns after every analysis. So far, I've only been successful at reducing the peak area (after prolonged cleaning with H2O/ACN gradients).
Any ideas?
-
- Posts: 2175
- Joined: Tue Aug 17, 2004 7:59 pm
Does this broad peak elute only during gradient runs?
Thanks,
DR

DR

-
- Posts: 480
- Joined: Fri Sep 03, 2004 7:57 am
How could ‘bacterial particles’ be eluted by mobile phase through the column? I think they tend to clog the column by their insoluble bodies than to create a band.
-
- Posts: 23
- Joined: Thu Aug 25, 2005 5:50 am
Dear DR and syx,
The broad peak, which I don't know whether is due to carryover from my sample, or something else entirely, elutes in both isocratic and gradient runs.
However, during gradient runs (increasing ACN), it elutes earlier.
Presumably if bacteria can cause high background absorbance, as DR said, the stuff the bacteria give off (metabolites? degradation products?) can absorb some UV.
The broad peak, which I don't know whether is due to carryover from my sample, or something else entirely, elutes in both isocratic and gradient runs.
However, during gradient runs (increasing ACN), it elutes earlier.
Presumably if bacteria can cause high background absorbance, as DR said, the stuff the bacteria give off (metabolites? degradation products?) can absorb some UV.
21 posts
Page 2 of 2
Who is online
In total there are 17 users online :: 1 registered, 0 hidden and 16 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: Google [Bot] and 16 guests
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: Google [Bot] and 16 guests
Latest Blog Posts from Separation Science
Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.
Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.
- Follow us on Twitter: @Sep_Science
- Follow us on Linkedin: Separation Science