negative peak for CO on methanizer/FID

Discussions about GC and other "gas phase" separation techniques.

2 posts Page 1 of 1
HI everybody
I'm running since many years an Agilent 6890 with FID to measure ambient air for CH4 and CO, (not CO2) through a methanizer converter. Time by time CO conversion efficiency seems to decrease, with 100ppb of CO giving a null signal (baseline flat) while I see, on a blank run (injecting N2 carrier gas or zero air), a deep negative peak exactly at the CO retention time. I've run several experiment to test for leaks, or column bleed but I can't find a solution: I still can not figure out the reason for the negative peak on the FID. Sample is injected by a sample loop; separation occur in a dual column system to avoid eluition of contaminant (heavier compounds) on main column (by Unibeads1S as a precolumn and Mol SIeve5A as analytical)
I've googled a lot on the forum and also on the web, but I've not find any report concerning a negative peak.
Could it be a matter of NiCat poisoned?
Any idea?
HI everybody...
DISCLAIMER: I have never used a methanizer on a GC/FID setup before. However, I was going to suggest that the catalyst may be the problem as you describe the "conversion rate" of CO goes down in time. I used to run Organic/Elemental Carbon analysis and this particular machine used a methanizer to reduce CO2 to CH4 to detect on an FID. The Methanizer catalyst usually had to be replaced annually.

I assume that as non-converted CO elutes and reaches the FID it is not detected it could interfere with your baseline signal as a "negative" drop.

Look here under "Disadvantages" and "Alternative Solutions."
https://en.wikipedia.org/wiki/Flame_ionization_detector
~Ty~
2 posts Page 1 of 1

Who is online

In total there are 64 users online :: 0 registered, 0 hidden and 64 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: No registered users and 64 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry