I have just read something interesting from the DryLab chromatography reference guide about how to handle baseline drift. It mentions adding a UV-absorber to the water to increase its absorbance and make it equal that of the organic.

This is a new approach to me. Does anyone practice this out there? We often get problems with our gradient baselines drifting and I was wondering if this truly is a way to counteract some of these problems. There is a mention of using salts and hydrophillic compounds. Anything specific in particular?

Anything you can do to get the RI of your A & B phases to match will minimize the gradient baseline effect.

I have never bothered with it.
Thanks,
DR
Image

Rob, the two references to the technique cited in Snyder and Kirkland's Practical HPLC Method Development book date back to 1982-83. The book states that "The matching . . . can be done conveniently by trial and error.", but then the authors are primarily theoreticians. :wink: I don't think I've ever met anyone who does this in "real life", and I can suggest two reasons:
- data systems generally handle a reasonable amount of linear drift quite gracefully
- by the time you get down to a sensitivity where the typical amount of drift is likely to be a problem, its magniture is usually far eclipsed by garbage peaks on the gradient (usually coming from CRUD in the A solvent).

As far as candidate additives go, a neutral salt like sodium nitrate would probably be a good choice for "UV absorbing and unretained". The catch is it that doesn't take much to provide the requisite few mAU.
-- Tom Jupille
LC Resources / Separation Science Associates
tjupille@lcresources.com
+ 1 (925) 297-5374
3 posts Page 1 of 1

Who is online

In total there are 16 users online :: 3 registered, 0 hidden and 13 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Ahrefs [Bot], Amazon [Bot], Google [Bot] and 13 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry