Hi all,
I have a specific problem whereby a 1.5 ml sample injection into a 5 ml loop @ 3.0 ml/min on a 100 x 10 mm column yields a retention time of 12 min with no problems yet the same injection on a 100 x 4.6 mm column @ 0.6 ml/min causes air to be injected onto the column and hence meaningless results. Could this be a back pressure-related problem?

Thanks

Here is my five cents of worth.

Just to make everything clear. You use the same system set-up, and load 1,5 mL sample using a 5 mL loop and where the only difference is the column dimension and the flowrate. If so, you have several effects playing with you. Now, if I've done my calculations in a correct manner.

1. on the 100x10, you load a fairly large sample volume (approx. 19 vol-% of the total column volume) using a sample loop that has almost the same volume as the void volume of the semi-prep column. At your reasonable flowrate it takes 1.67 min to elute the sample from your loop, giving a t'0 of 3.2 minutes. Your analyte elutes at 12 minutes, thus with a retention factor of 2,7.

2. Then you switch to a 100x4.6 mm column, where the total column volume is 1.66 mL and the void volume is 1.0 mL. You load 1.5 mL approx. 90 vol-% of the total column volume using a sample loop that has 5 times larger volume compared to the void volume of the analytical column. Now, as the flowrate you employ is scaled from the 10 mm id, your analyte (with a retention factor of 2,7) should elute around 37 minutes. In my mind, the extra-column effects in this second setup should be so severe that I can't really image that you would be able to get any results at all. 3. Only the band-broadening effects inside the sample loop should be so severe that it would prohibit data retrieval. It takes more than 8 minutes to empty the loop and I guess the diffusion must be rather substantial.

You must scale down the amount and volume loaded onto the analytical column to minimise the extra-column band-broadening effects and prevent volume overload too.

My personal rule of thumb which is also my reccomendation to other chromatographers:
Inject upto 1% of the column volume with full chromatographic efficency.
Inject 10 % of the column volume and achieve only 20 % of the full chromatographic efficency.

I have some slides available if it is of interest, otherwise it can be found in our HILIC guide.
Merck SeQuant AB
www.sequant.com
2 posts Page 1 of 1

Who is online

In total there are 13 users online :: 1 registered, 0 hidden and 12 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Amazon [Bot] and 12 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry