TCD of hydrogen in argon carrier
Does anyone have experience with the analysis of gas mixtures containing hydrogen in argon carrier gas?
I'm seeing a strange effect when constructing calibrations curves from either (1) injecting different volumes of pure hydrogen, or (2) injecting the same volume of mixtures containing different %'s hydrogen with balance air. For the same molar quantity of hydrogen, the H2 peak area becomes much smaller in the H2/air mixtures (becomes extreme below 40-50% H2) than in the smaller volume of pure H2. For example, the peak area of 22% H2 is about 100x smaller than the same molar quantity of 100% H2. These areas are reproducible from injection to injection and all the data fit together to form decent, nearly linear curves.
The only way I could think of to explain this effect is local concentration variation of hydrogen, because the 'slug' of hydrogen exiting the column in a dilute mixture will be more spread out than in the pure sample, therefore heavier in argon, with a reduced thermal conductivity. But even this ridiculous explanation (with empirical calculations of k of the mixture) doesn't seem to explain the huge difference.
Thanks for any suggestions.
Historically, H2 analysis is done with nitrogen carrier.
I believe your experience demonstrates why that is the case, although I never knew why argon was not a good solution.
Thanks for sharing your difficulties with Argon carrier performing H2 analysis. It has been educational for many of us on the Forum.
best wishes,
In total there are 22 users online :: 1 registered, 0 hidden and 21 guests (based on users active over the past 5 minutes)
Users browsing this forum: Ahrefs [Bot] and 21 guests