mobile phase vs sample solvent
Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.
5 posts
Page 1 of 1
I am attempting to analyze a DDT derivative solution using a C18 µBondapack column. My mobile phase is 55% ACN/45% water. My standards are prepared in acetonitrile and my samples are in a MeOH/water with 0.05 M sodium sulphate. I have no trouble seeing the peaks in my standards, but can no longer see the peaks in the samples. Both are at approximately the same concentration. I was previously able to see peaks in the samples using a spherisorb C18 column but was forced to switch to a µBondapack due to resolution issues. Could the difference in solvents be the cause? The injection volume is 5 µl, so I didn't expect a problem. Any other possible causes?
-
- Posts: 2916
- Joined: Mon Aug 30, 2004 10:19 pm
I can't see a good reason why this would be the case. Have you checked the sample vial?
-
- Posts: 2
- Joined: Sat Oct 09, 2004 6:21 pm
I haven't checked the vial itself but have injected several samples with the same result. It appears as though the compound(s) are being retained on the column, but I don't have an explanation (or solution) as to why this is happening?
-
- Posts: 2916
- Joined: Mon Aug 30, 2004 10:19 pm
Let me verify if I understood you correctly: you inject the sample onto the uBondapak C18 column in a mix of methanol and dilute sodium sulphate and you do not see any peaks. You inject onto the same column with the same mobile phase a sample of about the same concentration dissolved in acetonitrile and you see peaks.
If this is correct, I suspect that the issue is related one way or another to the injection. If you indeed inject sample from these vials, then maybe the issue is an adsorption of the sample to the vials or to the septa or realted issues.
If this is correct, I suspect that the issue is related one way or another to the injection. If you indeed inject sample from these vials, then maybe the issue is an adsorption of the sample to the vials or to the septa or realted issues.
-
- Posts: 586
- Joined: Mon Oct 04, 2004 3:00 am
There are deactivated sample vials available that are designed primarily for pesticide analysis. Your samples may be in a signithe ficantly different matrix than the standards, especially if you are analyzing extracts from biological materials. I would tend to suspect the samples are binding to the walls of sample vials, and that is why you are not getting a good response.
5 posts
Page 1 of 1
Who is online
In total there are 33 users online :: 1 registered, 0 hidden and 32 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: Ahrefs [Bot] and 32 guests
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: Ahrefs [Bot] and 32 guests
Latest Blog Posts from Separation Science
Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.
Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.
- Follow us on Twitter: @Sep_Science
- Follow us on Linkedin: Separation Science