Are you keeping the cholestane concentration constant while varying the cholesterol concentration?Thanks for that bit of information!
When you obtained different fractions of analyte cholesterol, did you have have to change the response factor you used? In the work I have been doing it seems as if there is a changing response factor depending on the ratio of cholesterol to cholestane being injected into the GC.
Also, i have been doing my injections "as is" into the GC. When I attempted the TMS in DMF I got a split cholesterol peak as if only a portion of it derivatized. I doubled the amount of reagents I used and got a similar result. Any thoughts on this?
Thanks in advance
Steve
I checked, from back in early 1990s, and we took up in pyridine, not DMF. GC: single sharp peak for cholestane, and single sharp peak for silylated cholesterol. So here's some test information from then:So we added cholestane early, did the extractions, then took up in dimethylformamide and made trimethylsilyl derivative of the cholesterol