% RSD in GC

Discussions about GC and other "gas phase" separation techniques.

3 posts Page 1 of 1
Dear All,
Is there any basis on which we keep % RSD limit of "Not more than 15%" in most of the residual solvent content methods?? :roll:
Am I missing something here? I always thought that calculation of %RSD was only to give you "warm and fuzzy" feeling that your instrument was performing as you expect. RSD's of more than you expect indicate a problem with your equipment. You can set what is "acceptable" wherever you want. If you're tied to some regulatory body, then you have to live with what they say. For residual solvents, if it's 10 ppm or 8.5 ppm or 11.5 and it doesn't matter, then 15% is a good limit. If it matters if it's 13 ppm then an RSD that is 30% is too large. You won't have a good idea that you're instrument is reporting a result that is "accurate enough" for the purpose of the analysis.
Dear All,
Is there any basis on which we keep % RSD limit of "Not more than 15%" in most of the residual solvent content methods?? :roll:
it's a compromise between cost and how fit for purpose the results are.

Peter
Peter Apps
3 posts Page 1 of 1

Who is online

In total there are 53 users online :: 0 registered, 0 hidden and 53 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: No registered users and 53 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry