I can't just switch the solvent with progressive gradient solution (like starting with 90% THF go to 70% THF and so on until I completely switch to DMF)?
-
- Posts: 2916
- Joined: Mon Aug 30, 2004 10:19 pm
Unfortunately, PEA is correct. These polymeric packings swell and shrink during the conversion from one solvent to another. This is why there are columns available that are shipped in THF and others that are shipped in DMF. There is a real risk that the columns that you are converting yourself may not survive this procedure. Please check the C/U manual!
-
- Posts: 16
- Joined: Wed Feb 22, 2006 1:54 pm
and if I switch to DMAC like PEA suggested? Is it the same problem?
-
- Posts: 16
- Joined: Wed Feb 22, 2006 1:54 pm
Which solvents can I use with column shipped with THF because I don't want to change all the column bank...
-
- Posts: 2916
- Joined: Mon Aug 30, 2004 10:19 pm
The THF columns can be converted to HFIP or to more non-polar solvents like methylene chloride. I do not have all relevant information here, but the C/U manual that came with the columns should have this information, including conversion protocols.
However, all things considered, if your polymer is soluble in DMF, I definitely recommend to use the DMF columns with the DMF. You may have a hard time finding a more non-polar solvent that does as good a job as the DMF does. You are going into the wrong direction.
However, all things considered, if your polymer is soluble in DMF, I definitely recommend to use the DMF columns with the DMF. You may have a hard time finding a more non-polar solvent that does as good a job as the DMF does. You are going into the wrong direction.
-
- Posts: 16
- Joined: Wed Feb 22, 2006 1:54 pm
Thanks to all for your collaboration, you have answered all questions.
-
- Posts: 16
- Joined: Wed Feb 22, 2006 1:54 pm
I'm back again with another question:
I need a second opinion on this: Could I run my THF-shipped columns with a mobile phase of 50% THF / 50% DMF because some of my latex are soluble in THF and some in DMF. I talked with a Waters technician and told me there's no problem as long as both solvents are miscible.
I need a second opinion on this: Could I run my THF-shipped columns with a mobile phase of 50% THF / 50% DMF because some of my latex are soluble in THF and some in DMF. I talked with a Waters technician and told me there's no problem as long as both solvents are miscible.
22 posts
Page 2 of 2
Who is online
In total there are 19 users online :: 1 registered, 0 hidden and 18 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: Google [Bot] and 18 guests
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: Google [Bot] and 18 guests
Latest Blog Posts from Separation Science
Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.
Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.
- Follow us on Twitter: @Sep_Science
- Follow us on Linkedin: Separation Science