I am running digested DNA (deoxyribonucleotides) on a Supelco LC-18-DB column with a guard column. As a reference I have been using dNTP's that are treated to resemble the digested DNA.
I use an isocratic run of 0.03M NH4H2PO4 with 2.5% methanol. In between each sample I wash with 66% methanol and 34% water.
I now seem to get split peaks, and broad peaks with a shoulder on every single peak.
Longer washing periods didn't seem to help so I upped the wash to 100% methanol each time, and then re-equilibrated my column with the running buffer. The next sample worked brilliantly, however it the broad and shouldered peak returned. 100% methnanol washes are no longer effective! I have tried very long wash periods, temporarily reversing the flow through the column and lowering the injection volume (only 10ul to begin with). The sample is mixed with the mobile phase prior to injection, so it shouldn't be caused by differences with the injection solution.
Is it possible that the column is no good? It has seen a fair few hours of work now.
I am stumped as to what would have caused this after running smoothly for some time. If anyone has any suggestions I am more than willing to give them a try.
Many thanks,