serap g » Fri Feb 24, 2006 8:09 am
Hello Einar
Thank you your answer. I don`t have a problem with amoxicillin but there is interference near the peak of clavulanic acid. My extraction method ;
To a tube ,500 ul plasma sample ,1500 ul ( 0.1 M HCI/ Acetonitrile; 1/9) were added. The mixture was vortex-mixed on a vortex mixer for 30 seconds.The mixture was then centrifuged at 5000 g 15 min. After centrifuged the supernatant was transferred a new tube. After evaporation to dryness at 350 C under a stream of nitrogen, the dried extracts werw resolved in 200 ul in water.
Mobile phase 0.01 M potassium dihiydrogen phosphate buffer adjusted to a pH 2.5 with orto phosphoric acid and acetonitrile ( 98/2)
Column Ace C8 ( 150 mm x 4.6mm x 5um) , room temperature .
I tried with SPE( C18) to separated interferences from clavulanic acid . but not separeted.
I don`t know, what can I do ???