Agilent 7890A problems

Discussions about GC and other "gas phase" separation techniques.

17 posts Page 2 of 2
Honestly I can't tell you why we are running such a high flow. I work off of Standard Operating Procedures and the current version says to run at such a high flow rate. It was also validated at that flow rate so that can not be changed.
Two things puzzle me;

why ask us to improve your procedure if you are unable to change anything that you are doing ? - to improve performance you have to change something in the procedure,

if the method was validated at 10 ml/min on a 6890 changing to a new model or instrument requires a re-validation anyway, so why not include a sensible flow rate in the new method ?

Peter
Peter Apps
dear peter

if you change flow rate no need to re validate method

if change column type or instrument then you need to verification not validation
( some parameter of validation not all )
17 posts Page 2 of 2

Who is online

In total there are 61 users online :: 2 registered, 0 hidden and 59 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Google [Bot], Google Adsense [Bot] and 59 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry