Hi
Just started using an ELSD microbore system, having problems with obtaining a baseline with flow rates of a 150 microlitres/min. I am using milli-q, methanol and formic acid for the solvent system.
I am new to this equipment so any suggestions will be more than greatful
Cheers
Sarah

A new ELSD takes time to blow the cobwebs out. The gas supply lines need to be very clean; if you don't have a sub-micron filter, get one. Also, the ELSD can tell you things about your water system that the conductivity meter won't tell you; if you have an alternative water source, see if that makes a difference. Finally, double-check all the advice in the manufacturer's instruction manual; there are some non-obvious things about the installation and setup of some models.
Mark Tracy
Senior Chemist
Dionex Corp.
2 posts Page 1 of 1

Who is online

In total there are 9 users online :: 2 registered, 0 hidden and 7 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Bing [Bot], Google [Bot] and 7 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry