Hi,

I am currently using the cleavable ICAT from ABI for comparative analysis of a bacterial pathogen. I'm not sure if this is a problem for anyone but it sure gives me werid results.

The last step of the labelling reaction requires cleavage of the biotin tag using TFA followed by complete drying down using speedvac to remove the TFA??. I realised the dried peptide seems to be hard to dissolve and it tend to give a few broad UV/TIC peaks during LC MS runs. Did anyone using ICAT observe this and what are your solutions?

Thanks in advance!

Ching Seng Ang

I think this is a so complicated reagent that you have to address this question to the people at ABI. I actually don't beleive that many here know what "icat" stands for.
------------------------
Merck SeQuant AB
http://www.sequant.com

Csang,

Normally you shouldn't have any problems resolubilizing your peptides. At that point you should had a very clean sample containing mainly cysteine containing peptides.

What might have happened is that your digestion in the earlier steps did not work very well so you still have some proteins at that point that are difficult to resolubilise. That would explain also your broad peaks.

Also after digestion you might want to clean up your sample with SPE before doing the rest of the chemistry.

Some more questions... do you use the ABI reagents or do you make your own TFA solution (if yes, you shouldn't as you do not know what else they might have added or what is the exact concentration of TFA they use). Finally, after we cleave, we dilute 10 fold before we dry down with the speedvac.

Hope the above helps,

Kostas

Dear Kostas,

Thank you for your suggestions. I'm currently using the reagents provided by ABi except the solubilisation reagent (using 6M urea, 0.05%SDS in 50mM Tris) as the recovery of protein seems better after acetone precipitation as compared to just 0.1% SDS alone.

I have thought of doing a SPE before all the clean ups, but after considering the low amount of protein (200ug total) I thought I would give it a miss. Did you carry out SPE before the SCX and Avidin steps?

The other thing I'm not sure about is the trypsin provided, whether it is modified since it only indicate TPCK treated. I think I will drop ABI an email to check on this.

Thaks again and best regards

Ching Seng
4 posts Page 1 of 1

Who is online

In total there are 30 users online :: 2 registered, 0 hidden and 28 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Ahrefs [Bot], Amazon [Bot] and 28 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry