We're experiencing some problems with an HPLC assay method for morphine sulphate. The retention time for the major peak changes from approx. 6.5 min to 13 min during a 10-15 samples run. We have also seen the opposite, i.e. from 13 min to 6.5 min.
We have tested different HPLC systems (Waters Alliance 2695 with Empower) with similar results.
We allow the HPLC to mix channel A and B of the mobile phases. However, we will try to pre-mix the mobile phase (I don't think this will help since we experienced similar problems on several systems).
Parameters:
Pre-column: LiChrocart 4-4, LiChrospher 100 RP-18, 5 um
Column: LiChrocart 250-4, LiChrospher 100 RP-18, 5 um
Pump flow: 1.0 ml/min
Mobile phase A: Acetonitrile
Mobile phase B: Ammonium acetate buffer (containing acetic acid, triethylamine and 1-heptanesulfonate monohydrate, pH 4.85)
Mobile phase ratio: A:B, 10:90
Injection volume 75 ul
Column temp.: 30°C
Degasser: On
Wavelength: 285 nm
The samples are prepared in water pH adjusted to 3.6 with phosphoric acid.
Approximate sample conc. is 0.096 mg/ml.
All solutions used have been vaccuum filtered.
My questions: Could this problem arise from anything but a bad column? Have I missed something?
I would really appreciate your input on this matter.
/Peps