analysis of ethylacetoacetate
hi
we are using HP-1 (30m0.53mm1.5µm) column for analysing ethylacetoacetate in GC FID. In this we are observing peak tailing and an another peak. In wax column the same one behaves differently , showing broad peak. give suggestion to get good peak.
mahesh
india
EAA
Use a high phenyl containing phase in your capillary column (VOCOL ?) and use non-protonic solvents if possible.
Keep your temperatures are low as possible.
good luck with a difficult analysis.
Mahs,
This has been covered in previous posts by me - see archives - and good luck. It breaks down to several peaks and also a peak which co-elutes with ethanol on DB624 type column (unless you use <160degC injection temp.). Peak likely to be ethanol.
I get 2 main peaks and fronting of peaks with DB5 type column at 250degC injection temperature.
Please share your experiences if you get any good peak shapes with this compound.
WK
In total there are 43 users online :: 1 registered, 0 hidden and 42 guests (based on users active over the past 5 minutes)
Users browsing this forum: Google [Bot] and 42 guests