Merging Chromatographic Peaks

Discussions about GC and other "gas phase" separation techniques.

3 posts Page 1 of 1
Hello,

I'm currentlly running a method for the quantitative determination of alendronate using a FLD detector. It seems that the peak of alendronate merges with a subsequent peak whose intensity remains constant. In most cases this merging results in a shoulder or extensive tailing of alendronate. Is there a way i can perform a split between these two peaks and are there any criteria for accepting the integrated in such mannner chromatogram?

Can you give more details on your method? Otherwise, is a bit difficult to give an opinion... :shock:

Looking at your method, you might be suffering either column degradation, since your pH is too high for "normal" silica based columns, or alendroate is partially ionized them generating a splitted peak.

Amines are known to create such problems. I would recommend you to try a Phenomenex Gemini C18 or similar column and increase your pH.
3 posts Page 1 of 1

Who is online

In total there are 17 users online :: 0 registered, 0 hidden and 17 guests (based on users active over the past 5 minutes)
Most users ever online was 5108 on Wed Nov 05, 2025 8:51 pm

Users browsing this forum: No registered users and 17 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry