Less ozone diffrence in SCD355
Discussions about GC and other "gas phase" separation techniques.
2 posts
Page 1 of 1
My GC SCD355 showing less ozone diffrence between ON and OFF postion (0.1 off and 0.2ON) I try to run 5 PPM H2S it showing only 0.5 ppm . Pressurw in detecor unit is 2.5 torr .
-
- Posts: 75
- Joined: Thu Aug 15, 2013 7:44 pm
Sounds like you probably need to replace the ozone generator.
Take the left side panel off thedetector box and then the smaller panel hiding the ozone generator. If the ozone generator is very discolored (dark brown/black or yellow), this is another indicator it needs to be replaced.
Take the left side panel off thedetector box and then the smaller panel hiding the ozone generator. If the ozone generator is very discolored (dark brown/black or yellow), this is another indicator it needs to be replaced.
2 posts
Page 1 of 1
Who is online
In total there are 67 users online :: 2 registered, 0 hidden and 65 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: Google [Bot], Google Adsense [Bot] and 65 guests
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: Google [Bot], Google Adsense [Bot] and 65 guests
Latest Blog Posts from Separation Science
Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.
Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.
- Follow us on Twitter: @Sep_Science
- Follow us on Linkedin: Separation Science