What pH should I maintain in my mobile phase when my compound is having 3 pKa values at 2,7 and 9?
I know that in reverse phase separation, a neutral compound will enhance retentivity in the column than an ioinised form.

I have successfully done trifunctional aromatic acids that are similar to your description with 0.15% phosphoric acid as water component.
Bill Tindall

The first question should be, in which pH range will you get adequate resolution of your compounds. Having said that, the very acidic pH range suggested by Bill is one possibility. Another very good possibility is the middle between the first 2 pK values, at pH 4.5, if an acetate buffer is suitable for detection. You should get stable chromatography there, with few issues. In the alkaline pH range, the two closely related pK values of your compound will result in less stable chromatography, meaning a high sensitivity of retention to small changes in pH. If you go to pH 10 to 11, you will get stable chromatography again. We have used pyrrolidine for this range.
3 posts Page 1 of 1

Who is online

In total there are 25 users online :: 1 registered, 0 hidden and 24 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Ahrefs [Bot] and 24 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry